jackaubrey77 ®   12-May-2020 11:38

ICES Cooperative Research Report No. 326. Calibration of acoustic instruments


Year: 2015
Language: english
Author: L. T. Hufnagle, X. L. Burton
Genre: Manual
Publisher: International Council for the Exploration of the Sea
Edition: May 2015
Format: PDF
Quality: eBook
Pages count: 133
Description: Acoustic instrument calibration is fundamental to the quantitative use of its data for estimating aquatic resource abundance. Regular calibrations also allow instrument performance to be monitored to detect changes due to the environment or component dynamics, degradation, or failure.
This is the second ICES Cooperative Research Report (CRR) focussed on calibrations of acoustic instruments. The first, CRR No. 144 (Foote et al., 1987), was published during the era of analogue electronics more than a quarter of a century ago. Since then, not only has the acoustic equipment improved vastly with digital electronics and signal processing, but the techniques for applying them to studies of marine organisms have both advanced and diversified. Motivating, facilitating, and expediting these developments is the work of the Fisheries Acoustics, Science and Technology Working Group (WGFAST) of the International Council for the Exploration of the Sea (ICES).
CRR No. 144 guided the fisheries acoustics community to uniformly apply the sphere method to calibrate survey equipment, generally single-frequency, split-beam echo- sounders. Today, surveys of fishery resources are conducted using a large variety of acoustic instruments including, but not limited to, single-frequency, multifrequency, single-beam, split-beam, broad bandwidth, and multibeam echosounders; side-scan and scanning sonars; acoustic Doppler current profilers; and acoustic cameras. These instruments differ in the ways in which they function, are utilized, and the types of measurements they provide. In most cases, they also require different calibration techniques for optimizing the accuracy and characterizing the precision of the measurements.
With technological innovation proceeding at an ever faster pace, the challenge to create a comprehensive and practical guide to calibrating acoustic instruments is formidable. Obviously, not all acoustic instrumentation and methods are addressed here. The ones that are addressed are in various states of maturity. Therefore, the practical aims of this CRR are to document (i) acoustic instruments currently used in fisheries research and surveys, (ii) theoretical principles of calibrating these instruments, and (iii) methods currently being practiced for a selection of commonly used instruments.

Contents

Screenshots

Calibration of acoustic instruments [2015]

Download [4 KB]

Thank U
Reply
Arcan   12-May-2020 12:35
jackaubrey77, From Rules:
8.6 The title should contain not more than two authors.
Fix please.
2
Reply
refresh list

Similar releases

Measurement Techniques, Sensors and Platforms - Various authors [2009, PDF]
EIVA NaviScan v.9.7 - EIVA [2022]
Advanced Metric Wave Radar - Jianqi Wu [2020, PDF]
EIVA NaviScan v.9.10.0 - EIVA [2026]
The Principles of Celestial Navigation - Adel Mostafa [2019, PDF]
EIVA NaviEdit v.9.2.0 - EIVA [2026]
High-resolution Seafloor Survey and Applications - Ziyin Wu, Fanlin Yang et al. [2021, PDF]
A High-Rate Virtual Instrument of Marine Vehicle Motions for Underwater Navigation and Ocean Remote…
Electric Machines: Modeling, Condition Monitoring, and Fault Diagnosis - Hamid A. Toliyat, Subhasis…
Dynamics Modeling and Maneuvering of Autonomous Underwater Vehicles - Theory and Practice - Xiang…